Multifunction DAQ
How to connect AIGND in case of multi differential measure with PXI6221
I need to make 8 differential measures with the PXI6221 of floating source. Some has low impedance and some has high impedance.
On the figure below I understand that inputs are multiplexed but it is the same for AIGND ?
Can I plug them as follow for 2 differential measures of source0=(S0+, S0-) and source1=(S1+, S1-) :
Floating source terminal     |     chanel name     |     pin number
___________________________________________________
S0+                                                 AI0                          68
S0-                                                 AI8                           34
S0-                                                 AIGND                     67
S1+                                                AI1                           33
S1-                                                 AI9                           66
S1-                                                AIGND                      32
Will S0- and S1- not be in short ?
Benoît Chantepie
(3,211 Views)
All Ai GND are connected together.
You can plug for 2 differentials measures as you did.
There will be no short circuit between S0- and S1-, both pins can be at the same location (understand same potential), there is no problem with that.
Regards
(3,201 Views)
Ok thanks,
So whatever the internal impedance of my floating source, I have to add a resistor between all AI- and AIGND to don't modify the behavior of my circuit.
Benoît Chantepie
(3,196 Views)
This is not a mandatory, if the impedance you measure is very low (<100 ohm) no resistor are required.
http://www.ni.com/pdf/manuals/371022k.pdf (pages 4-15)
Actually we add these bias resistors to avoid the source to float out of the maximum working voltage range, so to be sure the circuit works.
(3,193 Views)