Hello drj,
1. The flashing LED is normal.
2. The way you have proposed to count will not work because of the functionality of the counter. Please see this excerpt from the USB-6006 User Manual:
You can configure PFI 0 as a source for a gated invertor counter input edge
count task. In this mode, falling-edge events are counted using a 32-bit
counter.
What this means is that you can only increment the count in the counter on a falling edge of a signal. From what I understand that you are trying to do, you would be better off to use a counter's gate to set up a pause trigger, which is functionality not supported by your device. Instead, you should probably use a digital input to start a software timer when the input changes. You can see the LabVIEW Example Finder (Help >> Find Examples) for examples of how to do digital input with your device. In particular, navigate to Hardware Input and Output >> DAQmx >> Digital Measurements >> Read Dig Chan.vi
3. As long as your 6009 is connected to your computer, you should be able to read +5 or 2.5 V on terminals 30 and 31. If you are not getting this voltage, try connecting to a different USB port or even a different computer in case it is a problem with the port you are currently connected to.
Please let us know if you have additional questions.
Hope this helps!
Laura