Multifunction DAQ
Floating Source on USB-6356
Would there be a way to add a 1M input impedance to the USB-6356 device so that an oscilloscope would not have to be connected
Message 11 of 14
(380 Views)
Accepted by topic author rocksolidsr
The best option is to get the charge converter from the sensor manufacturer, if you could not do so, you can try adding the 1M load like below.
Santhosh
Soliton Technologies
Message 12 of 14
(373 Views)
Thank you, yes I have contacted the manufacturer for a quote on the charge converter, but in the mean time I will give that a shot and see how it goes.
Thanks!
Message 13 of 14
(370 Views)
I was able to try that solution and it is working just like when connected to the scope, so that will work until I can get a charge converter!
Thanks again!
Message 14 of 14
(365 Views)
