03-25-2020 03:43 AM
Hello,
I am having problems running tasks with channels connected to card TB-4300, which is mounted on PXIe-4300. If the channels are run in task with the acquisition mode as One Sample (on Demand), they seem to run. If the acquisition mode is changed to Continuous Samples, I get the error -50152.
Do anyone know who to run (or configure channels) on TB-4300 to be able to run on Continuous samples mode?
Thank you,
Laura
03-26-2020 11:43 AM
Hi Laura,
You should be able to acquire continuous samples. I don't have the physical hardware, but I have just created a simulated PXIe 4300 in MAX and run it in continuous sample mode.
If you haven't done so already, I recommend going through this KB: https://knowledge.ni.com/KnowledgeArticleDetails?id=kA00Z000000kFQGSA2&l=en-GB for troubleshooting steps.
Thanks,
Dan
04-02-2020 01:29 AM
Hello,
Thank you for your help. I have tried it and I can also simulate and run on continuous samples. I haven't fixed the problem yet but that link has been very helpful.
03-16-2021 02:54 PM
Hi Laura,
I'm having the same problem. I did what the help link suggested and it still doesn't work.
Were you able to fix the problem?
03-25-2021 08:30 AM
I was able to solve the problem by downloading and running the BIOS Compatibility software, and then resetting the NI-Max configuration data. I have a PXIe-1071 chassis with a PXIe-8360 controller, PXIe-4330 Bridge measurement card with the terminal block attachment (TB-4330).
The relevant information and instructions are on these pages:
I hope this helps anyone else who runs into this issue!