11-16-2023 09:17 AM
Dear all,
I am trying to connect an Hamilton pH probe to the NI-9208 module.
The pH probe is connected to a transmitter (Arc Wi 2G Adapter BT) and the wiring diagram is this one:
So, I have done the connection like that:
Arc 2G Wifi |
NI9208 |
1 |
AI0 |
2 |
COM |
3 |
AI1 |
4 |
COM |
7 and 8 are connected to a 24VDC
But it is not workin.... What did I do wrong?
Thanks a lot!!!!
Solved! Go to Solution.
11-16-2023 09:58 AM
As per your transmitter it needs a power source for the 4-20mA channels.
You must connect as per this diagram (NI-9208 will be your PCS),
Arc 2G Wifi |
NI9208 |
1 |
+24V |
2 |
AI0 |
3 |
+24V |
4 |
AI1 |
11-16-2023 10:11 AM
Thanks a lot Santosh, I'm quite a beginner in all that 🙂
Actually, that's the first thing I've tried, but it seems I did it wrong.
Also I need to connect the COM of the NI-9208 to the GND of the power source, right?
Thank you
11-16-2023 11:13 AM - edited 11-16-2023 11:16 AM
Yes, all three - sensor, transmitter and NI-9208 must share the same GND i.e., COM.
Depending on wiring constraints, this also works,
Arc 2G Wifi |
NI9208 |
1 |
VSUP |
2 |
AI0 |
3 |
VSUP |
4 |
AI1 |
+24V |
VSUP |
GND |
COM |
07-17-2024 08:04 AM
Dear all,
That's me again since I have some problem in connecting a Brooks Mass Flow Controller to the same NI-9208 and a NI-9266 for AO.
Please find below the connections for the MFC.
Here is what I did:
MFC | Card | Connection |
1 | NI-9266 | COM0 |
4 | NI-9208 | AI1 |
5 | + | |
7 | NI-9266 | AO0 |
9 | - | |
10 | NI-9208 | COM-17 |
Is this right?
Or should I connect as for the Hamilton probes like that (as COM-17 is already connected to '-')?
MFC | Card | Connection |
1 | NI-9266 | COM0 |
4 | NI-9208 | + |
5 | + | |
7 | NI-9266 | AO0 |
9 | COM-17 | |
10 | NI-9208 | AI-1 |
Connecting this stuff is definitely not my forte 😉
Thanks a lot,
Iforire
07-17-2024 09:04 AM
@Iforire wrote:
Dear all,
That's me again since I have some problem in connecting a Brooks Mass Flow Controller to the same NI-9208 and a NI-9266 for AO.
Please find below the connections for the MFC.
Here is what I did:
MFC Card Connection 1 NI-9266 COM0 4 NI-9208 AI1 5 + 7 NI-9266 AO0 9 - 10 NI-9208 COM-17 Is this right?
Or should I connect as for the Hamilton probes like that (as COM-17 is already connected to '-')?
MFC Card Connection 1 NI-9266 COM0 4 NI-9208 + 5 + 7 NI-9266 AO0 9 COM-17 10 NI-9208 AI-1 Connecting this stuff is definitely not my forte 😉
Thanks a lot,
Iforire
Please create a new topic so it is easy to find in future.