02-02-2010 09:19 AM
I need the specification for the sampling rate (internal clock) accuracy specification for the 446X series. I need to know how much the clock skew can be introduced perferrably with temperature info too. Where can I find this information?
Thank you.
Solved! Go to Solution.
02-02-2010 11:21 PM
Hi,
The NI 446X Specifications has this information. This is on page 14 which states the following:
Accuracy ......................................... ±20 ppm, over operating temperature range
The operating temperature range for this device is shown on page 15 of the manual as well.
Regards,
Todd V