Choppy signal ("stairs") from USB-6002
I'm geting a sort of a choppy signal from USB6002. After unsucsessfully trying solving ground problems, I connected just a capacitor with a discharging resistor directrly to the input- the same thing- you can see the "stairs". I also used elementary data readout, without the buffer, still the same thing (the time unit is 100 ms, 200 s total in x axis) Anyone may help?signal.png
import nidaqmx
import time
with nidaqmx.Task() as task:
task.ai_channels.add_ai_voltage_chan("Dev1/ai0")
d=[]
for i in range(2000):
d.append((task.read()))
time.sleep(0.1)
import pylab as plt
f=plt.figure()
a=f.add_subplot(1,1,1)
a.plot(d)
plt.show()
(1,122 Views)
Some additional details would be helpful,
  • Are you using USB6002 as AO or AI?
  • Graph shows 5mV steps - what is the capture device and configuration
    • Instrument is not evident from the python code
  • What instrument is generating the signal that is captured by the AI?
  • Circuit diagram of the test setup?
Santhosh
Soliton Technologies
(1,055 Views)
I'm using USB6002 as AI, both RSE or DIFF mode have the same behavior
The signal is generated by the pressure sensor, I replaced it however with the test circuit consisting of capacitor discharged through a resistor (see the circuit below)- I have the same steps.
The USB device is connected to the laptop working on battery- there are no other devices connected there. I dont see any options to configure USB 6002 through NI Daq software.
I used the input now in the DIFF mode- I have the same steps still...
The capacitor was charged by an external power supply, which was then disconected and taken away, as seen on the first plot. The steps ~5 mV are clearly seen on the second plot, which shows the same signal (zoomed view).
(1,021 Views)
Fig.1. Discharge of a 2200 uF capacitor
Fig. 2 Magnification of recorded signal, see the right scale [V]
 Fig 3. Test diagram
Fig 4. Test circuit
(1,019 Views)
