Can the USB-9211 Thermocouple Input device measure voltages ?

Can the USB-9211 be used to measure voltages ? It is designed for thermocouple measurements and its operating manual states there is a current source between its TC+ and TC- input terminals.

What amount of current is flowing ? Does it depend on the selected thermocouple type ? Can the current level be controlled by sotware ?

One application for me is to measure resistances of 1 ohm or less by measuring the voltage drop across the resistance induced by flowing a current through it. Might the 92111's current source be somehow used for this ?

What is its peak to peak input noise ? Its specs say 1 uV rms.

This is a delta-sigma converter. Is each successive measurement accurate or does some digital filtering have to be account for and thus it takes several readings for a signal change to become accurate ?
0 Kudos
Message 1 of 2
(2,469 Views)
The 9211 measures voltages between 0 and 80 mV only. The current source is only use to force a full scale input if an open thermocouple is connected and cannot be controlled by software. The input impedance of each channel between TC- and TC+ is 20M Ohm. However, the current that flows in the amplifier is very small, 50 nA.
The main problem here is the fact that the current source is not controlled by the user. This means that connecting a resistor of 1 Ohm between TC- and TC+ would not just produce a voltage that maybe due to external exitation, but the internal source should also be taken into account. Since the user has control over it, it would be a guess work to try to make sense of any voltage value read in.
its peak to peak input noise is 1 uV rms as indicated in the specs.
So as long as the measurements allow for enough settling time (12S/s), each one is accurate.
Message 2 of 2