Can a RTD be connected to a PXIe-6363?
We recently purchased a PXIe-6363 Multifunction module, and I was wondering if it is possible to use with a RTD with a constant current excitation
Thank you,
Greg A.
(1,974 Views)
You can use 6363's AI to measure the voltage across the RTD (with an external constant current excitation source).
You will be limited to the AI accuracy specs of the instrument and you can calculate the accuracy in terms of the temperature measured.
When I try to create a temperature task with the SCB-68A accessory, DAQmx suggests this connection.
Santhosh
Soliton Technologies
Only two ways to appreciate someone who spent their free time to reply/answer your question - give them Kudos or mark their reply as the answer/solution
(1,967 Views)
But is it possible for the PXIe-6363 to provide that constant current excitation?
Thank you
Greg A.
(1,936 Views)
Accepted by gacosta31
No, 6363 does not have an internal resource to provide the constant current excitation.
Only dedicated temperature input cards have such resource and not multifunctional X series cards.
PXIe-4357 would be the right option, it comes with an internal excitation current.
https://www.ni.com/en-us/support/model.pxie-4357.html
Santhosh
(1,921 Views)