06-17-2021 07:08 AM
I am using NI-USB 6259 bnc multifunctional DAQ device. There seems to be a problem by the ao pin values. It goes only from -2.5+2.5 volts when giving -10 to +10 in the test panel. This is the same for all 4 BNC outputs. I think for this reason the device was marked as broken by a colleague in our lab but when I took it it was working just fine for a while. Could anyone have any idea about the problem and possible solution?
07-07-2021 03:51 PM
That sounds like a problem I was having with my PXIe-6361's analog outputs. On mine, the outputs were limited to -3.3 to +3.3V even if I set the output to +10 or -10. I discovered that the board was also reporting the wrong ambient temperature in NI MAX. I ended up sending the board into NI for repair. I actually got it back this morning. The repair note says they changed the power supply and everything appears to work now.
07-13-2021 05:53 AM
Dear mj051,
Thanks for your reply. Indeed what you explained here is pretty much similar to my case. I was getting also reports on the wrong temperature plus I was suspicious that there might be something with the power supply as the one in use is not an original one. I think it is not worth sending my device for repair as it is quite old already. I will, however, try a dedicated power supply if I find it in the labs to see whether that solves the problem.
07-29-2021 09:10 AM
Do you tried with a dedicated power supply?
Did this solve the problem?
07-29-2021 09:16 AM
I sent my device to NI for repair. NI replaced the power supply on the PXIe-6361 board. I got my PXIe-6361 back and everything appears to work.