10-29-2008 11:51 AM
I am trying to use my AI-16XE-50 DAQCard to read accelerometers from a SCXI-1530 card. Installed Traditional NI-DAQ (Legacy) 7.4.4 for drivers. When I open labview and use the DAQ Assistant to obtain a signal for the DAQ it says "No supported devices found". I can see the AI-16XE-50 in Measurement and Automation Exporler. And when I go into the "Test Panels..." I see a signal but it does not seem to be correct because it is a flat signal that does not change as you ould expect (all of the channels).Thanks.
Ramin
Power Supply-SCXI-1001
Accelerometer Card-SCXI-1530
Accelerometers-Dytran Instruments 3055B2T
DAQCard-AI-16XE-50
OS-Windows XP
LabView 8.0
10-31-2008 11:31 AM
Hi Ramin,
The reason why the DAQ Assistant express VI says "No supported devices found" when you try to acquire data using your AI-16XE-50 DAQCard is that this device is not supported in DAQmx. Only DAQmx devices will work with the DAQ assistant. As for the test panels in Measurement & Automation Explorer not showing the correct voltages, there could be any number of causes for this. First, we need to verify that you have set up your SCXI-1001 and 1530 module correctly in MAX. If your hardware is set up correctly, your AI-16XE-50 should show up under "Traditional NI-DAQ (Legacy) Devices," and the icon should be green. Then, beneath your AI-16XE-50, you should see your SCXI-1001 chassis with a gray icon. When you expand the SCXI-1001 chassis, you should see all of the modules that are installed in that chassis, including your SCXI-1530. Is this the case with your setup?
Best wishes,
Wallace F.