09-12-2018 11:09 AM
Hi,
I am using a NI-9264 in a cDAQ-9189 chassis. I am trying to run the non-regeneration voltage output example. When I run this example, the output of the AO card does not update.
To clarify, If I start the vi with a sinewave of 10Hz and 1Vpp, it outputs the expected sinewave at 10Hz, 1Vpp etc. While the program is running, if I modify the parameters to a sinewave of 15Hz and 2Vpp, the output does not change. It remains at the 10Hz, 1Vpp.
Is this a hardware limitation where the output cannot be modified at runtime?
Thanks!
09-12-2018 09:25 PM
Maybe, this article helps you but is written in Japanese.... Couldn't find English version : (
Please translate with Google...
09-13-2018 07:10 AM
Hi,
Thanks for the info, however it doesnt really help solve the issue. My issue is regarding dynamically updating the output. Using the example, it runs at the correct frequency initially, but updating the parameters during run does not update the output.
09-30-2018 11:50 PM
After modifying the frequency and amplitude, have you pressed Enter on the keyboard or clicked somewhere else on the front panel so that the change you made to the array control gets registered by LabVIEW? There shouldn't be any hardware limitation for you to do this. After modifying the parameters, does the graph also not update with the new frequency and amplitude?
10-02-2018 10:53 AM
Hi there,
There definitely isn't a hardware limitation that is causing your unexpected behavior. There are a variety of programming factors that could be causing what you see. I would open up the block diagram on the example and check a few things:
1: You should see the "RegenMode" attribute being set to "Do Not Allow Regeneration" through a DAQmx property node somewhere in the example.
2: There should be a loop that is continuously generating fresh data that is being sent to the device using a DAQmx write vi.
3: The controls/variables that set the amplitude, frequency, phase, etc, that you wish to change dynamically must be INSIDE the loop, so that they are read every loop iteration and used when a new block of data is generated. Otherwise they will just be passed into the loop on its first execution and never checked again.
Hope this helps!
-Larry
10-02-2018 11:04 AM
Hi,
The issue here was relating to the Ethernet Chassis. It was a buffer issue. It was resolved by using the following: