6289 DAQ - AI_SENSE
Is it possible to wire a digital line into the AI_SENSE. That is wire PO.1 into AI_SENSE.
I ask because I need to Verify the levels for a control line on a DUT that is capable of producing +12V(+/-0.5V), +5V(+/0.1V), 0V and High Impedance (Hi-Z).
I was planning on wiring the PO.1 to AI_SENSE and use the AI_0 to measure the signal in NSRE mode to measure the +12V level and verify the High Impedance mode.
The NI6289 USB DAQ measure up to 10V on the AI lines and accepts upto 11V.  By applying a offset on AI_SENSE I should be able to read the 12V signal with the +5V offset provided by digital IO line.
The worries I have are (1) the stability of the 5V level on the Digital line (2) do I need to put a current limiting resistor between PO.1 and AI_Sense to limit the current should the High Impedance mode switching leave the DUT control level at 0V.
Any help would be appreciated.
(3,001 Views)
Accepted by topic author BeanBoy
If you read the spec carefully, you will see that the 11V input is referenced to COMMON, not the AI Sense.  So this won't work.
It seems like you need a voltage divider.  Two 100k resistors would work well to divide the voltage in half.  You can setup a DAQmx scale to account for this divider.
(2,988 Views)