I have a couple of questions:
1. What data acquisition device are you using?
2. What specific error code(s) are you receiving?
I hope you're having a great day!
08-20-2009 07:14 AM
08-21-2009 10:13 AM
Tom,
I assume you mean you are using the 9237. You are correct that this device does not a support a single sample read. This is not because of SignalExpress, however. The 9237 has a delta-sigma ADC that continuously samples with an oversample clock and has a built in integrator circuit that produces the measurement that is returned through software. The closest to single point you could get in your application would be to take a 2 point finite measurement and throw away one of your data points. That should be trivial to do in LabVIEW, but not so easy in SE.
Hope that helps,