Chandra,
The timeout error here is slightly misleading. The reason this is happening is caused by a few things. By default, the card to setup to transfer data when the onboard FIFO is full, and so your half buffer must be the same size or large than your FIFO unless you have regeneration turned on. Regeneration will cause your buffer to be copied as many times as possible into the FIFO until it is full. Your 6713 has an AO FIFO size of 16,384. So your buffer size must be twice this, so that your half buffer is >= 16384. Where is your data coming from at these 10 ms intervals, from a program or another device? This could affect what method would be the best solution.
If you want to keep working on this, let�s stick to one channel, I sent you an email, did you get
it?
Kevin R
Applications Engineer
National Instruments