I use the visa32.dll development tool
Declare Function viOpenDefaultRM Lib "VISA32.DLL" (ByRef sesn As Integer) As Integer
Declare Function viGetDefaultRM Lib "VISA32.DLL" Alias "#128" (ByRef sesn As Integer) As Integer
I find that if I use the "Keysight" driver. I can control the "GPIB0::20::INSTR" address of the instrument. But I use the "NI" driver. I will not be able to control the "GPIB0::20::INSTR" address of the instrument.
Why is this?
Are two visa32.dlls different?
I find the sizes of two visa32.dlls are different.
Solved! Go to Solution.
02-03-2020 02:22 AM
Hi
I have the same problem. Did you ever find a solution for this?
02-03-2020 08:23 AM
Use NI VISA or Agilent/Keysight VISA. Not both.
02-03-2020 08:53 AM
Ok, so it all happned when I updated to LabView 2019. The visa32.dll in the SysWOW64 folder updated to something that does not work with my old VB6 visa32.bas module. So is there another visa32.dll somewhere in a Keysight version?
02-03-2020 09:08 AM
You will have to find out how to use the Keysight VISA and not the NI VISA.