Hello all,
Setup: C# and the USB 6259 DAQ card. Two Analog output channels.
First question: What is the shortest time for data generation of the USB 6259 card? I cannot find the correct time in the specification
http://sine.ni.com/ds/app/doc/p/id/ds-20/lang/de
I need some simple C# samples where I can generate a square wave for example for 50 nano seconds (depeds on the minimum applicable time).
Thanks.
Regards.
09-05-2008 04:57 AM
Hi agent,
you should be able to find the answer to your first question in the link you posted. Among other things, it lists the Analog Output specs.
Maximum update rate
1 channel
2.86 MS/s
2 channels
2.00 MS/s
Timing accuracy
50 ppm of sample rate
Timing resolution
50 ns
Minimum output is 1 sample. So for 2 channels where the max output rate is 2MS/s, a sample width is 1/2M = 500ns with a resolution of 50ns.
As for your second question:
There are several examples for using the DAQmx driver with C# if you have Measurement Studio installed, you can find them here:
Select "Start"->"Run", enter %allusersprofile%
navigate to \Shared Documents\National Instruments\MStudioVS2005\DotNET\Examples\DAQmxWithUI\GlobalContinuousAO\cs
see also National Instruments\NI-DAQ\Examples\DotNET2.0
Let me know if you have further questions.
Best regards,
Peter