Learn calibration template error -1074396116

Highlighted
Hi,
 
When I used Learn calibration template.vi on an image 2500x2000 pixels with more than 5131 grid dots the vi return an error code :"-1074396116 Learn Calibration Template   Insufficient number of calibration feature points".
 
I don't find any limit in the documentation. Is anybody knows the source of this problem ?
 
Thank you
0 Kudos
Message 1 of 2
(1,427 Views)
Hi,

You've just reach a limit, which I admit isn't given, this should be corrected in next version. The solution as you may know is reducing the number of dots, by perhaps calibrating a smaller image, reducing the field of view (thus increasing the size of the dots and reducing the number) or using an ROI.
Hope this won't annoy you too much,

Regards

Richard Keromen
National Instruments France