01-16-2001 02:51 PM
10-27-2009 12:33 PM
I am trying to connect Tenney Environmental with VersaTenn III controller. I set up the connection parameters as follows: baud rate 9600, data bits 7, odd parity, stop bits 1, flow control Xon/Xoff. but when i try using the command
"? A1H" or "? SP1" it wont give any output. Is there anything i am missing?
10-27-2009 12:57 PM
10-28-2009 08:32 AM
I haven't used labVIEW. All i was trying to connect my PC to the controller that we have which is versatenn III via hyperterminal. I also verified connection settings, its says, Baud Rate: 9600, Data bits: 7, parity: odd, stop bits: 1, protocol: Xon/Xoff. I used RS232 cable to connect to the controller. I tested RS232 cable and its working fine. Any ideas?
10-28-2009 09:01 AM - edited 10-28-2009 09:02 AM
Are you using the correct type of serial cable? Does the manual specify null-modem or straight through?
Since you are not using LabVIEW, if you continue to have problems, you need to contact the vendor.
10-28-2009 09:26 AM
11-12-2009 02:06 PM
i am having same issue here, im using a gpib...
when i send the command like ? SP1, or any query command except from *idn?, i got
CONTROLLER COMM TIMEOUT
can somebody share info?
11-16-2009 02:52 PM
02-03-2011 09:20 AM
This is well over a year after the last post and I do hope that you guys have solved this, but I have a solution that should be posted here that would have helped me fix this a few weeks ago.
We have a Tenney chamber with a Versatenn III controller and I was constantly getting "CONTROLLER COMM TIMEOUT" messages any time I would attempt to connect with the controller. This message would be sent back over all 3 communication methods ( GPIB, RS232, and Ethernet)
After digging into how this system actually works I found that this typically is a Versatenn III setup issue, not the controlling PC.
First, open up the access panel that the Versatenn III is mounted in, if there is a circuit card attatched to the Controller that your GPIB, RS232 and Ethernet also connects to we're in business.
This is a Synergy488 communications card that acts as a translator between the Versatenn III controller and a PC. This card accepts commands from either GPIB, RS232, or Ethernet. To connect to this card using RS232 set your computers com1 settings to:
BAUD = 19200,
data bits = 8
stop bits = 1
parity = none
flow control = none
Be sure to use a straight through cable, a null modem cable will not work.
Hyperterminal works fine with these settings, use *IDN? to verify communication to the Synergy488 card.
If you're using GPIB, MAX will see the card and allow you to communicate. Some of you have been able to get the *IDN? querey to work.
Now that commuications with the Synergy488 card works, it's time to fix the CONTROLLER COMM TIMEOUT message.
On the Versatenn III, manually go to "SETUP", "SPECIAL FUNCTIONS" and set the VT3’s serial port parameters to these values: Baud to 9600, Data to 7O (7 bits data, Odd parity), COM to XON (Protocol = Xon/Xoff), and Log to NO. Data logging will disrupt communications.
communications with VT3 controller.
Commands sent to the controller should now work. Test is by sending "? C1" and you should get the temperature displayed on the VT3. The commands for this guy are a little wierd. For example; to set the temprature setpoint you must send "= SP1 xxxx" where xxxx is the temperature desired.
Tidal Engineering's program "SimpleComm" works great for manually sending commands since this is the same group that built the communication card. The manuals for "Linktenn32" and "Synergy488" is where I found the information.
02-03-2011 09:57 AM
I would also like to add the tech who originally set up our chamber didn't verify communications. We have the data sheet that has handwritten notes of the VT3 settings and he incorrectly set the special functions BAUD, and COMM. They were left to the default settings and must be changed if the chamber uses the Synergy card to communicate.