pci-6251 analog input set up

Solved!
Go to solution

Hi all,

Can I set the analog input of PCI-6251 in mix configuration (some in differential mode and some in single ended mode). Because in my set up the differential option is greyed out.

Thanks

dphan128

0 Kudos
Message 1 of 2
(2,621 Views)
Solution
Accepted by topic author dphan128
Yes you should be able to do that. You have 16 analog inputs. For differential you will use two of those per channel. See the DAQ M series user manual. You can program channels on an M Series device to acquire with different

ground references.To enable m

ultimode scanning in LabVIEW, use NI-DAQmx Create

Virtual Channel.vi of the NI-DAQmx API. You must use a new VI for

each channel or group of channels configured in a different input mode.

 

Remeber that for a differential that you would wire for example AI 0 to the + side and AI 8 to the - side. So now you can't use the AI 8 for single ended. AI 1 is related to AI 9 etc. up to AI 7, AI 16.

 

Hope this helps.

Now Using LabVIEW 2019SP1 and TestStand 2019