06-15-2021 10:09 AM
Hello everyone,
I am using a NI 9215 with BNC connector to acquire a voltage signal from Thorlabs DET02AFC photodetector, with a T4119 terminator connected between the NI card and the coax cable. Other ports on the 9215 are not in use. With this, I would like to measure light intensity from a M700F3 LED with 1.7 mW power, which theoretically should give me a signal of about 30 mV. But with the current setup, I get only about 3-4 mV at full LED power, measured via NI MAX.
To exclude hardware problems, I have tried the following:
- put a flashlight in front of the detector - I do not see any response at all, so I'm assuming that the LED is working correctly.
- varying LED power - I do see a response in the signal - a step increase or decrease when the LED brightness changes.
Could this be due to impedance matching between the LabView card and the detector? In the 9215 specs, I only found a cryptic (at least for me) information on the impedance:
NI 9215 with BNC (Between
any two AI- terminals)
200 kΩ
Any idea what could cause a 10x loss in measured voltage? I would be grateful for any suggestions.
06-16-2021 02:31 AM - edited 06-16-2021 02:32 AM
Hi agnef,
the datasheet says this:
due to the 100kOhm pulldown resistors connected to each AI- terminal (see page 5 of the datasheet).
Did you check the sensor signal using a simple handheld DMM?
06-16-2021 04:18 AM - edited 06-16-2021 04:19 AM
I just did, with the recommended T4119 terminator I also read 0.003 V. So I guess it cannot be due to NI card or LabView settings. If I remove it, I get a 13 V reading at maximum LED brightness..
06-16-2021 04:31 AM
Hi agnef,
Are you sure this T4119 still provides that 50Ohm resistance?
Did you verify this value with your handheld DMM, too?
06-16-2021 08:50 AM
Check the 12V battery in the DET02AFC, if it is dead or out of spec that may explain you lack of output voltage.
-AK2DM
06-16-2021 10:11 AM
So yes, the T4119 gives me 50 Ohms, I just double-checked.
And yes, I have checked the battery with the multimeter as well, it reads 12.45 V so I guess that is also ok.