Other notes:
NOTE 1:
I have attempted to use the Input Range values as the scaled values. For this load cell these correspond to
Fx= -50 to +50
Fy= -50 to +50
Mz = -200 to +200
When I input these LabVIEW (or at least the DAQ Assistant application) gives me an error code (-200007) that indicates "AI.Min" is incorrect. It then gives me a suggested maximum and minimum value for the input range. For the above these suggestions are as follows:
Fx = -30 to +30 (approx)
Fy = -29 to +29 (approx)
Mz = -36 to + 36 (approx)
I find this puzzling and do not know where or how "AI.Min" comes to these values.
NOTE 2:
Assuming the load cell is full bridge (and this has been confirmed with the vendor) I go in and simply set up DAQ Assistant as a simple Analog Voltage input. In the DAQ Assistant slot I select the three physical channels and when I go to run test I receive the same error code as in NOTE 1, above, but this time the Input Range values are suggested to be -0.005 to +0.005 volts. Again, I am puzzled as to how the "AI.Min" vi comes to this value.
NOTE 3: The values in NOTE 2 are especially puzzling as the system requires a 10 VDC excitation. I pre-set all appropriate jumpers on the SCXI board to meet this, per the SCXI User's manual.
NOTE 4: When I change the Input Range to the suggested range I get three readouts from each channel: all are set at the maximum of that range and do not change when I load or jog the load cell.
NOTE 5: An interesting day in the lab!!