05-06-2018 08:17 AM
Hello
I am looking for the option to change the compliance of the Keithley 2401 in labview.
How I can do this?
05-06-2018 10:52 AM
If you use the driver found here -
http://sine.ni.com/apps/utf8/niid_web_display.model_page?p_model_id=1583
Then when you configure the sense options one of the parameters is to set the compliance.
The SCPI command is something like..
:SENS:CURR:PROT 10E-3
To set a 10mA compliance when sourcing voltage or..
:SENS:VOLT:PROT 10
To set a 10V compliance when sourcing current.
Read the manual to get the limits of the compliance right if you are using a fixed range measurement.
05-07-2018 02:25 AM
I thank you very much
Unfortunately I did not find driver for the configuration of the sense options.
Can you help me with this?
05-07-2018 08:24 AM
If you download the driver, look at the VI tree and select the VI "Configure Output.vi" then one of the outputs is the compliance level.