Hi, All
I am using Agilent 532XX Series driver to read data from a frequency counter. In the low level subVi : \Public\Data\Low Level\Fetch Measurement (Single).vi, "Scan From String" is used to convert the data.
however, if I check the error of this function, it always show me a error with current format string "%.;%g;". after I changed it to "%;%g", I don't get the error anymore.
If anyone use this funtion, please check the error out of that fuction and see what you get.
Regards
04-09-2014 06:09 PM
I don't have the instrument, so I can't reproduce it completely, but I've done some research and if there is an error I don't think it's in the string conversion itself. I say this because:
From what I can gather, a typical output string would be the following: +4.27150000E-03
I took just a "scan from string" node and gave it that string as a string input and %.;%g; as the format specifier string. This yielded a numeric value of 0.0042715
So, as far as actually finding the root of the problem, there could be a few things going on. My first thought is some form corruption causing the data received to not be valid for conversion. There are two ways of testing this:
Open up a test panel in MAX and perform a loopback test, as described here: http://www.ni.com/white-paper/3450/en/#toc4
-or-
Install and open NI IO Trace and view the bus traffic as detailed here: http://digital.ni.com/public.nsf/allkb/282C5D41E2BA04F2862574BA007803B9
If the loopback tests works properly, go ahead and run the IO trace and post the results here.
04-09-2014 06:21 PM
04-10-2014 08:35 AM
Just in case you were wondering about any of the syntax in the format specifier string other than that period, this document explains it in detail:
http://zone.ni.com/reference/en-XX/help/371361J-01/lvconcepts/format_specifier_syntax/
04-10-2014 10:53 AM
@Dennis_Knutson wrote:
The period is there to force the conversion to use a period as the decimal separator instead of whatever the local setting is. Simple as that.
So, since the silly boys over at Agilent went and allowed the FORMat:DATA command. what you have is a bug and you should use this vi only when FORM:DATA? returns "ASCii,15"
Luckilly a *RST will put the device in that default condition.