do i synchronize two PCMCIA data acquisition card (NI DAQCard-6024E's)for continuous analog input in labview 6.0.2. Using example vi "Synch_E_PFI.vi", i get error -10802 after about 9.7 sec or i get error 10685. desired scan rate for 20 chan is 20-25/sec.Hello there! I think I found a possible source for your errors. In looking at your VI, you set up the first data acquisition board correctly, in that you route the AI scan start signal to PFI7. However, on the second board, I would NOT recommend doing th clocking as you have done it. Rather, I would get rid of the clock config VI altogether. For your Trigger config VI, set the trigger type to "Digital Trigger A", the mode to "On".
Now wire PFI7 from one card to PFI0 of the second card. PFI0 on the second card is the default for "AI Start Trigger", which is what you want.
Next, instead of the AI Control VI to start one of your cards, use the AI Start VI like you did in the first, and set your acquisition rate. Now the card that is waiting for
the trigger starts first, and is waiting. You then start your other board, and it should trigger the other, and both will start at the same time (you have it wired correctly now!).
Give that a whirl!
You're getting some errors now because one of your boards hasn't started (because it hasn't been triggered) at all, and there's no data to read.
Mark