‎02-17-2009 07:47 AM
I am using the fluke "getting started vi" and the error says <ERR> Fluke 45 Initialize
The ID Query failed. This may mean that
you selected the wrong instrument or
your instrument did not respond.
You may also be using a model that is
not officially supported by this driver.
If you are sure that you have selected
the correct instrument and it is
responding, try disabling the ID Query.
I have disabled the ID Query and I still get the same message.I have wired up the serial connector in the conventional way ,is there possibly an alternative wiring config.
Solved! Go to Solution.
‎02-17-2009 09:06 AM
‎02-18-2009 04:59 AM
I have gone into max & measure opened visa test panel
with the following result:-
read
count 1024 return count 4 return status 0
write
buffer *IDN?\n return count 6 return status 0
‎02-18-2009 06:50 AM
‎02-18-2009 07:44 AM
Dennis,
I am not that familiar with the visa test panel so could you please be more specific how I should use it correctly.
frank
‎02-18-2009 09:46 AM
Dennis,
To be more specific where do i find the text returned option.
frank
‎02-18-2009 10:09 AM
The text returned is the text that is in the 'Buffer' box on the Read tab.
You can also test communication with a terminal emulation program such as Hyperterminal. Just type the command that requests the instrument to send back the identification information (i.e. *IDN?).
‎02-19-2009 05:46 AM
Dennis,
Thank you for your help.The problem has been solved , the meter had been configured to even parity instead of none (parity bit 0).
frank