Hello!
Some colleagues face an annoying problem, and they asked me to help them. Their system: a cDAQ-9184 Ethernet chassis with two modules: NI9217 and NI 9211. OS: Win10 32bit. LabVIEW 2016, 32.
Every time they reboot the PC, the connection is lost to the modules. It seems the ethernet connection is OK between the chassis and the PC, but the chassis does not recognize the two modules. They were using DHCP settings, I changed this to static IP connection, but as I expected, this did not solve this problem.
Actually this is not a huge issue, but quite annoying: every time, if you need to reboot the PC, then you must open MAX, and click on the modules with the mouse. Magic: the chassis recognizes the modules now...
Can this annoying problem to be fixed somehow? See two screenshots below, the first what we see after reboot in MAX (our LV application of course reports error about the missing cDAQ modules), the second is when we click on the module entries in MAX, and they magically get recognized...