Where can I go to learn about scaling voltage readings from different AI devices?

Solved!
Go to solution
Ex. I have a Vaisala HMP 60 RHT sensor. How do I know what to scale the voltage readings to get the correct RH? I have seen values (such as 20 in the linear range for the Vaisala HMT 100), but I don't know where it comes from or why it is chosen. In case it matters, I am using several AI devices (thermocouples, MFC's, RHT's) on an NI-6343 USB board. I'd really like to learn more, but am finding information very tough/convoluted. Thank you for any/all help and direction.
0 Kudos
Message 1 of 3
(2,424 Views)
Solution
Accepted by DHagan

Hi DHagan,

 

I think that this article might help you.

 

Best,

 

Carmen C

0 Kudos
Message 2 of 3
(2,399 Views)

Thank you for the input/help. I think what I was looking for was that if you have a 5V output (for the Vaisala HMP60), then the slope would be 20 so that the zero voltage reading corresponds to 0% RH and 5V output corresponds to 100% RH. Hopefully this is right and their is no drift.