01-06-2018 12:18 AM
Tektronix TDS 3032C Oscilloscope uses Ethernet connectivity. In NI MAX and LabVIEW VISA IO resource, the TDS 3032C is detecting. But unable to communicate in LabVIEW. Instrument driver, all NI Drivers installed.But same problem
01-06-2018 01:49 PM
Can you communicate with the instrument in MAX? Can you share your code?
01-08-2018 04:44 AM
Unable to communicate in NI MAX. I have attached the LabVIEW coding (This coding is download from NI, but unable to communicate with TDS3032C)
01-08-2018 06:30 AM
@phoenix-karthik wrote:
Unable to communicate in NI MAX.
You really need to solve this issue first. This tells me you have a connection issue. What IP address is your computer using? What IP address is the oscilloscope set to? What subnets do you have these set to?
01-08-2018 07:51 AM
PC IP - 169.254.108.87
Instrument IP- 169.254.108.88
Subnet mask - 255.255.0.0
Port number - 80
I can able to ping to instrument IP.
01-08-2018 09:23 AM
What VISA name are you using? Did you try using the VISA discovery in MAX? Is the oscilloscope set up to accept commands through the Ethernet port (it sticks in my mind that you had to set this somewhere in the utility options)?
01-09-2018 10:59 PM
I have installed both NI VISA and TekVISA in my PC. In my application i have to communicate with three Instuments (DMM, AFG and Oscilloscope). When both VISA (NI and Tek) installed, I can able to communicate with AFG and DMM and When only TekVISA installed, I can able to communicate with Oscilloscope only. This conflict occurs.
01-10-2018 06:44 AM
Uninstall the TekVISA. It is not needed. I have talked to that same model with just the NI VISA over both GPIB and Ethernet without an issue.
Alternatively you could go into MAX, expand the tree on the left to get to My System->Software->NI VISA XX.X (XX.X = version you have installed). Then in the main window's tree go to General Settings->Passports. Make sure the NI-VISA Passport for Tulip is checked.
01-10-2018 12:08 PM
You say that the setting is for port 80. I'm pretty sure port 80 is for the device's internal web server, and if you're trying to send commands to that it won't work. Don't set it up as a raw socket, set it up as a VXI-11 LAN instrument.
01-11-2018 12:17 AM
But the instrument is not detecting in VXI-11 Auto detect of manual entry. Is there any Driver needs to be installed for VXI Connectivity?