12-30-2008 02:12 PM
I have a NI 9219 and an omega j type thermocouple,
When i connect the thermocouple to NI9219 terminal 4 and 6 ( positive and negative terminals) and open up Signal express I don't see the voltage change! is there something else that i need to do before i can see the measurments?
12-30-2008 04:06 PM
I'm using the USB9211, and had to get the max and min voltage set right before it would display. It had to be between 8.0e-3 and -8.0e-3 for us to get a good reading. Not sure if this will help, but if so, great!
Matt
01-02-2009 06:02 AM
Hi there
you have to use terminals 4 and 5 for thermocouples. See page 11 on the user manual: http://www.ni.com/pdf/manuals/372407a.pdf
01-02-2009 10:41 AM - edited 01-02-2009 10:41 AM
chrisger wrote:you have to use terminals 4 and 5 for thermocouples. See page 11 on the user manual: http://www.ni.com/pdf/manuals/372407a.pdf
That is correct. In my experience (the little there is), thermocouples are almost always in adjacent terminals