Hunter,
Thanks for the screenshot! The good news is that what you are experiencing is perfectly normal. The reason for this skew is that you must use a PXI-Trig line (or RTSI as it's referred to in Traditional DAQ) to start each board. The PXI Trig lines are wired in series from slot to slot, so there will be a time delay between each board as the pulse propagates down the line (resulting in a t0 difference). This is not a significant source of error, however, because every subsequent sample conversion is controlled by the Star Trigger which, as you know, has equal trace lengths between slot 2 and all of the other devices (unless you're using an 18-slot chassis). Even with the star trigger, the first board will make its conversion several nanoseconds before the others since it is the source of the pulse -- but given that the 4472 only samples at a rate up to 102.4KS/s, several nanoseconds is negligible.
You've done everything correctly.
Please let me know if you have any further questions!
Thanks again,
Alan L