Hi guys,

 

I have a load cell (full bridge) which has the following characteristics:

 

Min. Input Signal = 0

Max. Input Signal = 2500 kN = 2500000 Newtons (Compression)

Excitation of 10V is provided internally by the module

Sensitivity at 10V = 20019.8 microvolts

Input impedance = 1050 ohms

Output impedance = 1050 ohms

Output without load = -137.367 microvolts

 

I use DAQmx to model the load cell. The load cell certificate doesn't provide any details regarding scaling that needs to be entered in LabVIEW. I have calculated the following: 

 

Considering load cell supply is 10 Volts for max. signal of 2500 kN, and since sensitivity at 10V = 20.0198 mV, then

 

20.0198/10 = 2.00198 mV/V. So scaling information is: 

 

0            mV/V    -   0   Newtons

2.00198 mV/V    -   2500000   Newtons

 

LC.PNG

I am not sure If I need more values in the scaling information in DAQmx or these two values should simulate the load cell at its full capacity. Any thoughts should be much appreciated. 


Thanks. S

 

 
0 Kudos
Message 1 of 4
(2,623 Views)

When you use Scaling, you need to know several things.  One is the "model" of your instrument -- is the Voltage Out a linear function of the Stimulus In?  If so, then you can use linear scaling, which needs two parameters that I'll call "offset" and "gain".  Offset is the voltage when the input signal is 0, while Gain is the change in voltage corresponding to a known change in input.

 

Assuming that (a) your Load Cell is linear (according to the manufacturer) and (b) the Offset is 0 (which you can test by taking a reading with a Load of 0), your calculations for determining Gain seem correct.  But be careful -- if your device is designed to output 0-10V, and cannot (by design, for example) output -0.1V, then for all you know, an output of 0V with a load of 0 might be because the system wants to output -0.1V, but is "clipped" at 0V, a sign of a non-linearity in the Instrument and therefore a violation of the assumption for which you set up your scaling rules.  An easy test for this is to take an "unloaded" reading, a "one-penny" reading (put a penny on the Load Cell), and a two-penny reading.  Choose pennies with the same year, as the formulation can change.  The voltage change from 0 to 1 penny should be exactly (!?!) the same as the difference between 1 and two pennies -- if there is a significant deviation, add a 5-penny value, plot the 0, 1, 2, and 5 penny values, and see if a straight line goes through the points, but hits the Y axis at a negative number.

 

Bob Schor

0 Kudos
Message 2 of 4
(2,593 Views)

Thanks Bob. 

 

The load cell is linear yes, however, if i use the calculated scale and there is an offset from the load cell, this offset can be deleted in the post-processing analysis of the recorded data right?

 

Here is the attached certificate of the sensor for your convenience:

 

ca.PNG

 

Do I need to take anything else in consideration?

0 Kudos
Message 3 of 4
(2,563 Views)

This sounds OK.  However, I'm a scientist, by training, so I always believe in "doing an experiment" (like the Penny Weighing I mentioned) to demonstrate that things work as you think they should (or not).

 

Bob Schor

0 Kudos
Message 4 of 4
(2,545 Views)