PLease check out the following shipping example that includes Strain gauge nulling and shunt calibration for the SCXi 1520(Assuming this is the module that your 1314 connector block is being used with).
Either of the following VIs could be used as a template of howto implement calibration:
Cont Acq Strain Samples (With Calibration) - SCXI 1521.vi
Cont Acq Strain Samples (With Calibration).vi
These two vis Can be found under: .....\National Instruments\LabVIEW 8.6\examples\DAQmx\Analog In\Measure Strain.llb
This should be able to help you considerably!
04-17-2009 05:14 PM
04-17-2009 10:47 PM
Do you want your half bridge 1 setup to use the shunt cal resistor? In my opinion, the shunt cal probably isn't necessary, and it just confuses all the wiring.
Without the shunt cal, you can use just 2 wires from each gage, red and black. The red of the R4 gage goes to P+. The red of R3 and the black of R4 goes to S+ and it is actually best of those two are joined togther close to the gages and only one wire goes back to S+. The black of R3 goes to P-.
If you put in the shunt cal resistor, then take the white of R3 to one SCA and another wire whichever is unused (red R3, black R4, white R4) to the other SCA. Remember in the sketch, any wire in the schematic that has an RL resistor shown means it is a long wire that has an inherent resistance. In that drawing only 5 RL's are shown so you only need to use 5 total wires that have roughly equivalent length.