08-07-2011 11:21 AM
Dear All,
I have a system with two PCI 5122 digitizers (one master and the other slave) synchronized together via RTSI cable.
In particular, the digitizers share the 20 MHz clock on the RTSI-0 and the reference trigger generated by the first digitizer on the RTSI-1.
Now, I need to use such a trigger signal to activate an external device that uses a falling TTL and, therefore, I cannot directly use the PFI0 of the master digitizer.
As in the system there is also a PCI 6601 board, I would like to ask which is the best solution to route the trigger available on the RTSI-1 to the 6601 and invert its polarity.
Many thanks in advance.
Best,
Mario
08-08-2011 08:28 AM
Hi Mario,
you could export the 5122 trigger signal to a PFI line (using function niScope Export Signal.vi) and connect this PFI line to a PFI line of 6601.
You can use 6601 to generate a pulse with desired polarity using PFI as a trigger. You can consider DAQmx example Gen Dig Pulse - Dig Start.vi.
I hope this helps!
Bye!
Licia
08-08-2011 08:32 AM
Hi again,
just a note: for this issue you should post to High-Speed Digitizers or Counter/Timers boards.
Licia
08-08-2011 08:56 AM
Many thanks Licia for the advice.
Indeed, I was already using the PFI connector to make this route and I was wondering whether there was a more straightforward solution via the RTSI cable.
Best,
Mario