Read data of sartorius scales with RS232 interface

Hi Dennis,

Thanks again for your reply! Baud rate, parity etc. are correct (I can read manually back these settings from my device).

How can I check if I need a null modem or straight through cable?

How can I confirm that I send the carriage return, the line feed and the escape – how do I have to set this in my program?

Sorry for asking such perhaps stupid questions – so far I had seldom big LabVIEW problems, but since the RS232 interface is totally unknown to me, I really need detailed help.

0 Kudos
Message 11 of 26
(9,649 Views)
The manual for the instrument should tell you the type of cable. You already said you checked it. How could you check it if you don't know the correct type?

I have never used your terminal emulation program so I don't know it's specifics. At this point, you don't have s LabVIEW problem.
0 Kudos
Message 12 of 26
(9,644 Views)

You have not mentioned which model of balance you have. One manual I checked has a very nice cable diagram and complete specifications on the RS-232 protocol. It uses somewhat unusual factory settings: Odd parity and hardware handshake.

 

It is quite easy to verify the string being sent. Just wire a string indicator to the write buffer input to the VISA Write. Set the display style to '\' Codes Display or Hex Display.

 

Lynn

 

verify string.png

Message 13 of 26
(9,634 Views)

Thanks johnsold, I adapted my program as you suggested – but still no data.

The model is: Sartorius E 12000S (e.g. here is a short manual: http://www.sartorius-lab.net/uploads/1/0/6/2/10629561/e5500s_e8100p_e12000s_instructions.pdf). And yes, it uses odd parity (as set in my program). But I cannot judge if this is an unusual setting since this is my first RS232 interface (and at the moment I hope it will be the last).

Regarding the cable: In the manual is written which pin is used and if I understand my 25-pin (device connection) to 9-pin adaptor (at my USB adaptor) correctly, it should work fine.

 

Perhaps after some hours of sleep I will find with the support of you/the forum members a solution (perhaps the model number is of help for you).

Anyway, thank you so far for all your efforts!

0 Kudos
Message 14 of 26
(9,625 Views)
There is nothing in that manual that describes the serial connection or commands. What was the point in attaching it?

0 Kudos
Message 15 of 26
(9,617 Views)

Dear Dennis_Knutson,

I was not able to find a more detailed manual for this balance in the internet (I fear it is too old); I have the manuals in print form.

But I found a manual which is very similar to my one:

http://www.geminibv.nl/labware/sartorius-bp210s-balans/sartorius-basic-manual-eng.pdf 

 

E.g. the general specifications are on p. 55. Here it differs slightly from my manual:

Standard: I have additionally RS423

Handshake: I only have CTS and DTR (and nowhere in my manual is mentioned if and how this can be changed)

Transmission rates: my maximum is 9600 baud

Synchronization: I have 1 stop bit

Data output: I have 22 characters.

 

On p.60 you can find the data input formats and on p. 61 you can find the command ESC P CR LF which I have to use to get data of my balance.

 

The pin assignment is on p. 66, but this differs a lot from my one.

 

I will try to find a manual of my device in the internet. If I do not find one, I could scan in the relevant parts of my manual.

0 Kudos
Message 16 of 26
(9,593 Views)

Might be a stupid idea but have you tried to send an uppercase P instead of the lowercase p? If it is an old device, its interpreter may not account for case insensitive commands.

 

Other than that, going from your VI in your post 8 I don't see any other reason that it won't work if cabling and serial port settings are ok, which of course is hard to check without at least some basic RS-232 tester such as a breakout box with LEDs that show the physical states of the lines.

Rolf Kalbermatter
My Blog
Message 17 of 26
(9,584 Views)

Hi rolfk,

Thank you for this hint – I tried to use P and p but this has no effect.

But it is good to know that such an experienced user as you thinks that my vi should work.

 

Due to the responses here, I think at the moment that I do something wrong with the device itself. 

So, tomorrow I will meet a guy who has used this balance successfully together with a computer (decades ago ;-).

I will keep you up to date when I have new information!

0 Kudos
Message 18 of 26
(9,572 Views)

I usually use something like this for testing serial port connections. It shows with LEDs the logical state of each line and allows easy detection of cabling problems such as swapped data lines or wrong use of handshake lines. It also allows to rewire the lines individually so you can easily test various forms of wiring schemes other than the one-to-one connection. Another more simple solution looks like this here. It does show the state of the LEDs but doesn't allow to rewire the connections.

 

One inside joke is that RS-232 is an acronyme for "resolder pin 2 to 3 and vv."

Rolf Kalbermatter
My Blog
0 Kudos
Message 19 of 26
(9,565 Views)

The manual you posted clearly states that the commands are case sensitive (pg 61).

 

Have you tried the Automatic Data Output mode, also known as auto print? In that mode it should sned data continually. It would ease the troubleshooting process on your receiving end.

 

Lynn

Message 20 of 26
(9,536 Views)