Read Measurment from HP-3458A

I'm trying to read the current measurement of a HP-3458A Multimeter (DMM) using the LabView instrument driver. But, when I try to start the measurement appear a Query error code -1074000000, and the LabView read the multimeter current value as an error. Someone could help me with a solution of this issue?

Captura de tela 2022-05-20 133203.png

Captura de tela 2022-05-20 132418.png

0 Kudos
Message 1 of 3
(1,219 Views)

Replace the faulty GPIB cable.  One of the handshake lines is not right.

 

Your device is not clearing a buffer with the *RST command and responding with what appears to be a reading to SYST:ERR?  That can only be a hardware fault.


"Should be" isn't "Is" -Jay
0 Kudos
Message 2 of 3
(1,177 Views)

I changed the USB/GPIB interface and its working now. Thanks so much!