Hello,

 

 

 

I think i'm overlooking something simple but would like clarification on measuring RPM via a quad encoder.

 

I have created a DAQmx task to read quad encoder angular position. I have configured the DAQmx task to read angular position in Revs. Samples are read via continuously buffered samples at 1kHZ rate and reading 100samples at a time with no issue. 

 

 

To convert to RPM, I take the first (0) and last (99) sample from the 1D position array, calculate deltaRevs over the sampling period to get deltaRevs/deltaT and then convert to RPM. 

 

Since I'm reading 100samples at 1kHz, I'm believe I should be using a time period of 100ms. However, this produces a 1% error in RPM reading. It seems like I need to use 99ms as my time base to get the correct result. 


What am I missing? 

 

Regards,
Dave

0 Kudos
Message 1 of 3
(2,199 Views)
Solution
Accepted by DavidT

Hi David,

 

between sample #0 and sample #99 are 99 dt intervals…

(From sample #0 to sample #1 it takes 1× dt, from sample #0 to sample #10 it takes 10×dt, from sample #0 to sample #99 it takes 99×dt.)

 

Usually you compare sample #0 from current data block with sample #0 from previous data block and use your calculated 100ms per data block…

Best regards,
GerdW


using LV2016/2019/2021 on Win10/11+cRIO, TestStand2016/2019
Message 2 of 3
(2,197 Views)

I thought it was that simple. Smiley Embarassed

 

Thanks for the clarification!