PXI NI-4065 set 3 and half digit with 4 wires resistance on lab labview

Dear Sir 

    We ask to suggestion.

We try to set labview for control PXI NI4065 by item below

1. 4 wires resistance measurement for measure 0-200 ohm

2. Set labview with resulotion 3 and half digit.

 

Please suggestion that config is correct and stable for measurement or not.

0 Kudos
Message 1 of 2
(677 Views)

What is your definition of stable?

 

You can improve your measurement repeatability by increasing the number of averages. Lower digit resolution would result in faster measurements and probably noisy.

 

You can give a try if this configuration will satisfy by making measurements using Instrument Studio and looking at the statistics like Standard Deviation, min, max

Santhosh
Soliton Technologies