12-17-2021 01:15 AM
install visa to a new PC , in NI-MAX found the error as the photo show ,
is it need to update the visa version to the newest ?
12-19-2021 06:46 PM
could any engineer give some suggestion?
i did no setup the same IP address for each equipment. i will uninstall drivers and install one by one to check if this error happen again.
12-19-2021 10:07 PM
Have you installed any other visa drivers other than NI-VISA in your system for example TEK VISA?
It might happen if you have installed multiple VISA drivers.
12-19-2021 11:14 PM
thanks for your responsing.
i installed MOXA pci serial ports extendtion.
i have uninstalled all NI drivers and then install them again , it seems the error dispeared, i will double check again