03-14-2019 10:18 AM
Essence of the problem is that I'm having difficulties getting NI MAX to see an Agilent 34970.
I'm setting up a new computer to run a LabView runtime application. Thus far, I have done the following:
Install LabView runtime 15.0
NI-VISA runtime 15.0.0
NI-DAQmx runtime 15.0.1
NI-DAQmx 15.0.1
Keysight Connection Expert
I tried to download the agilent 34970 driver located here: http://sine.ni.com/apps/utf8/niid_web_display.download_page?p_id_guid=014F4CD7EF78136AE0440003BA7CCD..., but the directory doesn't exist that I'm supposed to copy the files to.
In device manager the Agilent shows up as a CP2102 USB to UART Bridge Controller.
Solved! Go to Solution.
03-14-2019 10:59 AM
Are you using a USB to serial adapter to connect to the serial port on your device? If so, have you installed drivers for the USB to serial adapter?
03-14-2019 03:50 PM
Yep that's all it was. Thank you!