NI 9211 Thermocouple Inaccurate Reading

Hello, 

 

I'm trying to build a simple VI that stores temperature versus time measurements from an Omega Type K thermocouple in a CSV file (please see attached). I am using an NI 9211 coupled with a NI cDAQ-9172 chassis (due to my older equipment, I am still using LabVIEW 2017).

 

The VI I have thus far seems to behave appropriately, but when I use my thermocouple to measure the temperature of an ice water bath, I get consistent erroneous readings of 50 degrees Celsius. I've verified that the thermocouple works with a 4 channel datalogging thermometer, so the problem lies with my program or the DAQs. 

 

In addition, I know the "Convert to Thermocouple Reading.VI" is not compatible with the NI 9211; is there any conversion I need to apply within my program to get the correct temperature reading?

 

Appreciate any insight!

0 Kudos
Message 1 of 15
(2,217 Views)

Your sample rate is set to 1kHz, the 9211 supports rates up to 14 S/s. TC's usually read much slower than other analog signals (at least in my experience) I usually just read 1 sample every 250 ms in a software timed loop. Start with that, you may want to also consider converting the express VI to proper Daqmx code. 

0 Kudos
Message 2 of 15
(2,206 Views)

Hello, 

 

Thanks for your quick response. I reduced the sample rate to 3 Hz, but I still observed the same 50 degree measurement of an ice water bath. The module is quite old, so maybe it needs to be manually calibrated.

0 Kudos
Message 3 of 15
(2,132 Views)

Hmm.. Maybe. The cDaq 9172 is listed as a legacy device. Can you open up MAX (Measurement and automation explorer)  and read correct temperature with the test panels there?

 

 

0 Kudos
Message 4 of 15
(2,122 Views)

No, I still encounter the same issue. 

0 Kudos
Message 5 of 15
(2,101 Views)

Not a good sign. Maybe time to open a ticket with NI or get new hardware. There may be a way to recalibrate it and get it working, but I don't know how.

0 Kudos
Message 6 of 15
(2,094 Views)

Thanks for your help!

0 Kudos
Message 7 of 15
(2,089 Views)

You tried different thermocouples?

 

I would suggest that the thermocouple itself is most likely duff.

Nick
0 Kudos
Message 8 of 15
(2,066 Views)

Hi Nick, 

 

I have three thermocouples and two NI 9211s. All port combinations with all three of my thermocouples yield the same result. 

0 Kudos
Message 9 of 15
(2,062 Views)

Rubbish.

Probably the device then. 

Can you verify the thermocouples on some other device maybe a hand-held meter?

Nick
0 Kudos
Message 10 of 15
(2,033 Views)