06-17-2019 10:58 AM
Thank you for the response.
I only have one Oven.
Idk why all the other GPIB addresses show up.
I tried Status?\r\n and Temp?\r\n.
The oven works on another computer that's running an older Windows OS.
The cables are not even close to 2m long.
I attached a picture with nothing attached to the GPIB-USB-HS
06-17-2019 02:30 PM
I also installed the latest drivers NI-VISA, 488 and DAQ
06-17-2019 03:34 PM
Time to start at the beginning. I suspect your problem is with the GPIB-USB setup and not the oven at all.
- What sort of GPIB-USB are you using? (NI, Keysight, etc.)
- Have you specifically installed the driver for the GPIB-USB? (If not an NI device then you need the driver software for it.)
- Do a scan for instruments in NI-MAX and show us the screen cap of the oven at the expected address. If it doesn't show up, then the GPIB-USB connection is your problem. Unplug it, re-install driver, reboot, plug it in and make sure it is properly configured.
- Next step is to make sure you can demonstrate that the GPIB-USB is working with another instrument. (Proving its the oven's fault.)
Once the GPIB-USB is working with another device, move on to debugging the oven. Hopefully you are disconnecting PC#1 before attaching PC#2. Only attach one PC at a time and try the commands mentioned before.
Craig
06-18-2019 12:33 PM
Thank you for the response.
I will try all these steps and get back to you asap.
06-20-2019 10:09 AM
Thanks all for the help.
It was the NI-VISA driver.
I had to download 16.0.
All works.
06-25-2019 01:38 PM
Hey, I asked in a new forum, but maybe its quicker if I ask you. Is it possible to use the PID function in LabVIEW for the GPIB device?