The Nexus has a RS232 interface. So you can set the appropriate amplification with LabVIEW. The technical documentation of the Nexus interface is more than 100pages .....
For the LabVIEW side you need the VISA driver installed to communicate via RS232. Search the NI driver network maybe there is already a B&K Nexus driver ....
12-01-2010 09:54 AM
The USB-9233 has a 24-bit analog-digital converter (one 1 bit is lost for bipolar; theoretically this means about 138 dB), but the datasheet defines a range of about 102 dB .
The input range is 5 V.
As the range is 102 dB the finest voltage is about 39 µV (microVolt).
Now, 39 µV (microVolt) is about 7,94E-04 Pa. This value, to the base level of 20e-6 Pa for 0 dB, defines 32 dB (linear).
For low noise measurements, about 32 dB is the effective noise floor of the 9233. If you use the Nexus signal conditioning to increase the gain, you can lower this but you run the risk of clipping if you don't get it right and have higher signals.