Keysight 34972A reduce number of switching processes on 34901A 20-Channel-Multiplexer

Dear all 🙂

 

I am using Keysight 34972A in combination with the 34901A 20-Channel-Multiplexer to measure voltages.

Right now I only measure one voltage. later there might follow more.

 

Everytime, I use the measure voltage block the device switches the relay. If I choose a fast measure rate, the relay will be end of life very soon.

Is there a way to measure one channel without the relay switching all the time? (actually it only has to switch the relay when I change the channel, right?)

If someone has an idea I would be very happy 🙂

 

thx 4 input

 

Kind regards

Anna ❤️

 

 

0 Kudos
Message 1 of 8
(2,529 Views)

Hi Anna,

 

so you made a VI to read a voltage measurement from your DMM.

And something in your VI is switching this relay.

 

And now we should tell you what to change in your VI - without being able to examine your VI?

Best regards,
GerdW


using LV2016/2019/2021 on Win10/11+cRIO, TestStand2016/2019
0 Kudos
Message 2 of 8
(2,508 Views)

Hi GerdW,

Sorry, you missunderstood me. I used the measure voltage block from the LabView-library of the manufacturer (Keysight).

Best Regards

Anna

0 Kudos
Message 3 of 8
(2,472 Views)

Anna,

If you read this except from the manual,

The following program segment shows how to use the MEASure?
command to make a measurement on one channel. This example
configures the instrument for dc voltage measurements, internally
triggers the instrument to scan one channel, and then sends the reading
to the instrument’s output buffer.
MEAS:VOLT:DC? 10,0.003,(@301)

This implies that the system will reconfigure itself with the next MEAS command/query.  However, If you use this command to set your conditions initially then by sending READ? for all the following readings, you will get the next DMM measurement without re-configuring the system.  It will not change any of the relay connections using the READ? query.

 

Also, here is the manufacturer's data except on the 34901A modules:

  • The Keysight 34970A Data Acquisition / Switch Units along with the 34901A modules use the G6SU-2-DC3 relays.
    • Omron is one of the manufacturers providing these relays and are specified as having:
      • minimum electrical dc duration of 100,000
      • minimum mechanical duration of 100,000,000

 

Help the Community (and future reviewers) by marking posts as follows:
If it helped - KUDOS
If it answers the issue - SOLUTION
0 Kudos
Message 4 of 8
(2,449 Views)

If you want to continuously measure/monitor ONE channel (instead of scanning a channel list) use the Monitor VI.

 

But just so you know we have been using these devices in our lab for probably 20 years and I have some 34901A MUX cards that are getting on 20 years old that still function as good as the day they we received them

 

Also maybe this is not the best instrument for your application.

 

The 34970A and 34972A are designed to take a large amounts of measurements at relatively slow but consistent intervals, not so much for constantly monitoring a single test point

 

========================
=== Engineer Ambiguously ===
========================
0 Kudos
Message 5 of 8
(2,437 Views)

An alternative SCPI command to the MEAS command/query would be:

CONF:VOLT:DC 10,0.003, (@111)
ROUT:MON:STAT ON
ROUT:MON (@111)
ROUT:MON:DATA?

Now continue to send the last command/query to receive new voltage readings when requested.  The issue with this method is if you change the channel being monitored by hand on the front of the device later in the process (causing the relay contacts to change), it will read the new active channel.  To avoid the accidental change, you can send the last 2 lines in a repeating fashion.

Help the Community (and future reviewers) by marking posts as follows:
If it helped - KUDOS
If it answers the issue - SOLUTION
0 Kudos
Message 6 of 8
(2,420 Views)

Use this VI to constantly monitor/measure one channel without opening and closing the relay every time

 

mon.png

 

You have to configure all your channels using the "advanced configuration" functions first!

 

And NO, the "EZ Config" vi's will not work with the monitor vi.


@RTSLVU wrote:

If you want to continuously measure/monitor ONE channel (instead of scanning a channel list) use the Monitor VI.

 

But just so you know we have been using these devices in our lab for probably 20 years and I have some 34901A MUX cards that are getting on 20 years old that still function as good as the day they we received them

 

Also maybe this is not the best instrument for your application.

 

The 34970A and 34972A are designed to take a large amounts of measurements at relatively slow but consistent intervals, not so much for constantly monitoring a single test point

 


 

========================
=== Engineer Ambiguously ===
========================
0 Kudos
Message 7 of 8
(2,412 Views)

Message 4 above should be saying "EXCERPT" and not "EXCEPT".  Gotta love some of the auto-correct functions.