Hello I have a problem using my Keysight 34461A. When I run my program multimeter says it's on remote mode but the program itself freezes on a intialize.vi block. I'm using labview 2018 and have correct drivers.

1.png

Bez tytułu1.png

Is there any solution?

Message 1 of 14
(3,780 Views)

May be it takes a while to initialize? you can step inside the Initialize VI during execution to see where exactly it freezes.

Santhosh
Soliton Technologies

New to the forum? Please read community guidelines and how to ask smart questions

Only two ways to appreciate someone who spent their free time to reply/answer your question - give them Kudos or mark their reply as the answer/solution.

Finding it hard to source NI hardware? Try NI Trading Post
0 Kudos
Message 2 of 14
(3,765 Views)

First thing to try is setting ID Query option to FALSE.  You may have to modify your Initialize VI if the string your instrument returns isn't handled correctly.

 

EDIT:  Of course, you should also verify your VISA resource is set correctly and physical connections are good with NI-MAX.

LabVIEW Pro Dev & Measurement Studio Pro (VS Pro) 2019
0 Kudos
Message 3 of 14
(3,725 Views)

@NIquist wrote:

First thing to try is setting ID Query option to FALSE.  You may have to modify your Initialize VI if the string your instrument returns isn't handled correctly.

 

EDIT:  Of course, you should also verify your VISA resource is set correctly and physical connections are good with NI-MAX.


Just an FYI that when you disable the ID query for HP/Agilent/Keysight, it will default to a very basic set of commands that may not even work for your instrument.  It depends on getting the ID to send the commands appropriate to your instrument.

Bill
CLD
(Mid-Level minion.)
My support system ensures that I don't look totally incompetent.
Proud to say that I've progressed beyond knowing just enough to be dangerous. I now know enough to know that I have no clue about anything at all.
Humble author of the CLAD Nugget.
Message 4 of 14
(3,712 Views)

That explains that "User Data" variable sent to the instrument I was wondering about.  The Initialize VI reads the model of the instrument and enumerates it into an integer.   Looks like in this case it just sets it to the simplest model in the line (the 34460A = 0) if you don't run the ID query.  Interesting.  Thanks for the info Bilko!

 

NIquist_0-1671744870250.png

 

LabVIEW Pro Dev & Measurement Studio Pro (VS Pro) 2019
0 Kudos
Message 5 of 14
(3,705 Views)

One small correction: It does NOT send that User Data to the instrument. A VISA session provides a storage attribute that allows the user (here the instrument driver) to store some arbitrary information that it can later query, in order to do device specific handling. On the VISA API level this is a pointer sized integer but LabVIEW has no pointers (and the closest thing to pointers, a DVR, didn’t exist for almost another 20 years, when VISA was invented and integrated in LabVIEW).

 

You can think of this like a private element in a LabVIEW class and the property node is the class property accessor. All VISA property nodes don’t directly correspond to a device element but rather to an attribute of the underlaying communication interface or a VISA internal session management attribute.

Rolf Kalbermatter
My Blog
Message 6 of 14
(3,641 Views)

@AdrianOstrycharczyk wrote:

Hello I have a problem using my Keysight 34461A. When I run my program multimeter says it's on remote mode but the program itself freezes on a intialize.vi block. I'm using labview 2018 and have correct drivers.

1.png

Bez tytułu1.png

Is there any solution?


Everything said by others hear is good information HOWEVER, it fails to address one potential issue. 

 

That DMM has a triggering subsystem.  If the device is actively working on measuring almost nothing is important to the device.  In fact, it will simply store most commands for later action until it is finished with its primary purpose of taking measurements.  The "*RST"  command will abort any active measurement.  

 

I would suppose that you are not taking any action to abort an ongoing measurement.  So, the instrument is simply too busy trying to do the important work you left it trying to finish.   Like a dog you placed on a stay command, your trivial distractions don't mean anything to the well trained device.

 

 


"Should be" isn't "Is" -Jay
0 Kudos
Message 7 of 14
(3,579 Views)

idquery.png

initialize34461a.png

initialize53220a.png

   

It freezes on attempting to reset and always ends on a timeout no matter how long the timeout is. I also tried using another Keysight instrument 53220a. The result was exactly the same. I'm running out of options. Is there any possibility that there's problem with my USB port?

Message 8 of 14
(3,548 Views)
  • Open the VISA test panel in MAX.
  • QUERY "*IDN?" Report exact string
  • QUERY "SYST:LANG?" Report exact string

How exactly are you connected?  You mentioned USB.  Is there a USB-GPIB adapter or, are you using a Virtual COM Port?  Is the instrument configured to use the correct bus?  If you are using a VCP Is the hardware driver installed?


"Should be" isn't "Is" -Jay
0 Kudos
Message 9 of 14
(3,526 Views)

I have a 34461A connected through USB and I get this for my *IDN? response:


Keysight\sTechnologies,34461A,MY53219439,A.02.14-02.40-02.14-
00.49-01-01\n

 

I would suggest that you close your VI and LabVIEW to make sure all VISA resources are really closed.  Then open the example called Agilent 3446X Series Acquire Single.vi since it is the simplest thing to run so you avoid any confusion.  Best to troubleshoot with known good examples and this works fine for me.

 

If it still fails the same way:

Open the RESET.VI since it not only sends a *RST, but also runs the DEFAULT.VI and then checks for errors where it may be stuck.  Turn on highlighting to see which VI is the actual culprit.  Then, open and debug that to see where the root cause is.

 

If all else fails, you can try manually resetting the instrument too (Press Shift > Run/Stop button) to clear any errors and reset to factory settings.  If all that doesn't get you going, you may have a hardware issue with your USB or cabling.  Obviously try the other USB ports on your PC with a known good cable.

 

 

 