06-25-2019 03:39 AM
Hello together,
i am using a Keithley 3706. In the Keithley slot is connected a measurement card 3720 with 3720-ST. ( The 3720-St is needed to connect thermocouples to the Keithley)
The issue is, if i set up the thermocouple type from K (K is Regular on Keithley 3706) to Type J. The Keithley does have a problem to hold the thermocouple type J.
In every loop he is jumping between type K and type J.
I upload a small example. I do not know where is my mistake.
Please Help me.
06-26-2019 08:06 AM
Hi Apotec,
make sure, that you do not set Type-J and reset in each iteration. Most probably there lies the mistake.
Try to remove the other VISA write functions one-by-one to check, which resets your instrument.
Best Regards,
NI TSE