How to connect Position sensor (Thorlab) to the NI DAQ (6343)?

Solved!
Go to solution

Hello,

 

Working on project where I have to read the position of the light beam using optical position sensor.

I use optical position sensor from thorlabs (PDQ30C) and for the data acquisition I use NI-DAQ(6343).

 

Now position sensor has 6 pin Hiros connection cable and now I need to connect the position sensor to the DAQ , but do now have idea how to connect and at what port of the DAQ I have to connect.

 

Position sensor :

https://www.thorlabs.de/newgrouppage9.cfm?objectgroup_id=4400&pn=PDQ30C

 

Link shows six pin connector  and how each pin gives different signal.

 

NI-DAQ 6343:

https://www.ni.com/documentation/en/multifunction-io-device/latest/usb-6343/pinout-bnc/

 

Could any one give me some idea how to make connection between position sensor and DAQ for data acquisition.

 

Thank you

0 Kudos
Message 1 of 7
(3,151 Views)

Looks like you need a +ve power supply line and a -ve power supply line, and then two Analog Input values to measure the X and Y values.

Connect pins 1 and 2 to your Analog Inputs of choice, and 5 to the common (GND) value for both the power supply/ies and DAQ.

Connect 4 and 6 to +- voltages respectively.

 

For others' reference, the datasheet can be found here: https://www.thorlabs.com/drawings/b9c5691259b472d2-9F6EB426-A265-47DD-59C2F13832EADE1A/PDQ30C-SpecSh...


GCentral
Message 2 of 7
(3,117 Views)

So in short,

I connect 1 and 2 pin to the analog input and 5 to the common.

Can I apply power supply from DAQ by connecting 4 and 6 pin to the one of the analog output port?

0 Kudos
Message 3 of 7
(2,973 Views)

In short, no.

The voltage levels on an analog output are suitable but the maximum current is much too low for your sensor.


GCentral
0 Kudos
Message 4 of 7
(2,968 Views)

I checked on specification of the DAQ (Analog output ) it shows output current +- 5mA

 

and limit for the sensor is 35mA

 

mi1111_0-1579600714801.png

 

 

mi1111_1-1579600732466.png

0 Kudos
Message 5 of 7
(2,961 Views)
Solution
Accepted by topic author mi1111

The DAQ value is a maximum. The sensor value is a requirement. You need at least that much.

 

You can if you need supply the positive value from the DAQ board via the +5V output, which can source 1A, but you have no negative supply.


GCentral
0 Kudos
Message 6 of 7
(2,956 Views)
Solution
Accepted by topic author mi1111

Hi mi,

 


@mi1111 wrote:

and limit for the sensor is 35mA


That's not a "limit", but a requirement!

Use an external power supply!

Best regards,