From Friday, April 19th (11:00 PM CDT) through Saturday, April 20th (2:00 PM CDT), 2024, ni.com will undergo system upgrades that may result in temporary service interruption.

We appreciate your patience as we improve our online experience.

LabVIEW

cancel
Showing results for 
Search instead for 
Did you mean: 

How many sample rate groups I can use in a 9189 chasis with 8 NI 9231

Solved!
Go to solution

I have a 9189 cDAQ chassis and several NI 9231 cDAQ modules. 

 

How many AI tasks with different sample rates can I have.

 

Thanks.

 

Ian 

0 Kudos
Message 1 of 6
(2,508 Views)
Solution
Accepted by topic author TulaRen

3


"Should be" isn't "Is" -Jay
0 Kudos
Message 2 of 6
(2,493 Views)

Thanks. That's what I remembered, but I just cannot find anything reference to this from ni.com.

 

Ian

0 Kudos
Message 3 of 6
(2,489 Views)

Its down deep in a cDAQ whitepaper


"Should be" isn't "Is" -Jay
0 Kudos
Message 4 of 6
(2,479 Views)

Hello Ian,

 

In this case you can actually only run 2 hardware timed analog input tasks with different sample rates because the 9231 is a DSA module. This is because the cDAQ 9189 is a Gen II chassis and Gen II chassis support a maximum of 2 synchronization pulses. This is called out here.

 

Regards,

Conrad

C Series IO R&D

Message 5 of 6
(2,471 Views)

Thank you for the correction and link


"Should be" isn't "Is" -Jay
0 Kudos
Message 6 of 6
(2,464 Views)