We need to read the measurements from our Fluke 45 over an GPIB to our PC ( Labview 5.1 ). The problem is that the things we see on our PC are different to what our Fluke 45 displayes. Where just using the VI's from the driver for the Fluke 45 ( from NI ). Can it be that there are some settings that we must change before we can start. Doesn't maybe anybody has the right VI's for us ! thanks in advance
Hi... What exactly are you seeing that is different? is it the measurement values? resolution of readings? If you can supply more details on the nature of your problem, then maybe we can find a solution... Cheers JB
the problem may be due to the setting of the decimal delimiter character: if on your PC the decimal delimiter is set to ',' then a conversion from Fluke's ASCII data to another format might give wrong results. GPIB instruments practically all use '.' as the delimiter and therefore driver VIs usually make a conversion based on this setting.
If ',' is set on your PC, change it to '.' (under country settings) and try again.
-Franz
Pauwels schrieb:
> We need to read the measurements from our Fluke 45 > over an GPIB to our PC ( Labview 5.1 ). > The problem is that the things we see on our PC are different to what > our Fluke 45 displayes. > Where just using the VI's from the driver for the Fluke 45 ( from NI > ). > Can it be that there are some settings that we must change before we > can start. > Doesn't maybe anybody has the right VI's for us ! > thanks in advance