Hi,

 

I'm having a problem with sampling below a rate of 1k with the NI USB-4431.

 

Possible reason(s):

Requested value is not a supported value for this property. The property value may be invalid because it conflicts with another property.

Property: SampClk.Rate
Requested Value:  100.000000
Valid Values Begin with:  1.0e3
Valid Values End with:  102.400000e3

 

Attached is my vi.

 

I've tried changing the wait but it makes no difference.

 

I'd really appreciate some help on this one please.

 

Thanks in advance,

 

AT12.

0 Kudos
Message 1 of 5
(2,593 Views)

Unable to open your vi. Can you post it in a lower version?

 

Also, what is the sample rate you tried? I assume you are using DAQmx timing function, or the DAQ Assistant.

 

As mentioned in the error message, the device accepts sample rates between 1K and 102.4K samples/sec.

 

Try with samples rates above 1k and things should work fine.

 

 

Regards
Freelance_LV
TestAutomation Consultant
Message 2 of 5
(2,588 Views)

Hi,

 

Thanks for replying. Smiley Happy

 

It does work fine higher than 1k but it is supposed to measure lower than this:

 

Four channels on the USB-4431 are TEDSanalog input channels that can sample from 1 S/s up to a full rate of 102.4 kS/s.

 

I need to measure at <1k S/s. I am using the DAAQ assistant.

 

I've Saved for Previous Version: 2009 - attached.

 

Thanks,

 

AT12

 

0 Kudos
Message 3 of 5
(2,583 Views)
Solution
Accepted by topic author AT12

Hi

 

http://www.ni.com/pdf/products/us/cat_usb4431.pdf

 

As per the specificaitons in the above document, the sampling rate is from 1kS/s to 102.4kS/s

 

Regards
Freelance_LV
TestAutomation Consultant
Message 4 of 5
(2,579 Views)

Oh right! I got my information from the overview here:

 

http://sine.ni.com/nips/cds/view/p/lang/en/nid/206676